Minutes of the 8th Liquid Scintillation Users Forum Module 16, NPL 7 May National Nuclear Laboratory. Babcock Marine-Rosyth Ltd
|
|
- Thomas Bruce
- 5 years ago
- Views:
Transcription
1 Minutes of the 8th Liquid Scintillation Users Forum Module 16, 7 May 2009 Attendees: Simon Jerome Arzu Arinc (Chairperson) (Secretary) Ben Ahern Rob Allott Eleanor Bakhshandeiar Andrew Boschier Keith Bradshaw Jane Caborn Philippe Cassette Lorraine Currivan John Davies Alan DuSautoy Ronald Edler Maria Garcia-Miranda Chris Gilligan Terry Gingell Chris Hardy Peter Harris Robert Huggett Marty Johnson Steven Judge David Lambert Etienne Letellier Olivia Marsden Andy Pearce Gary Prior Charlie Stewart Simon Temple Klaus Thieme Richard Wilson G E Healthcare Environment Agency Veterinary Laboratories Agency AMEC NUCLEAR UK National Nuclear Laboratory CEA RPIR Beta-Gamma Training PerkinElmer DSTL Babcock Marine-Rosyth Ltd Zinsser Analytic USAFSAM Magnox Veterinary Laboratories Agency UKAEA AWE Scientifics Ltd DSRL DPM Solutions Ltd Nuclitec GmbH Westlakes Scientific Consulting Apologies: Pete Burgess Nuvia
2 1. Chairman s welcome / previous minutes and actions arising Simon Jerome introduced himself and welcomed members to the eighth meeting. A straw poll showed which nuclides the participants were interested in and measured routinely: H-3, C-14, P-32, S-35, Cl-36, Fe-55, Ni-63, Kr-85, Sr-90, Tc-99, Pm-147 and Pu-241. A couple of participants mentioned using Cerenkov, gross beta and gross alpha counting. The minutes of the seventh meeting were accepted as a true record. No actions were raised from the previous minutes. 2. Invited talk: Determination of Carbon-14 in environmental level, solid reference materials Jane Caborn, NNL There is no suitable reference material for C-14 in more complex matrices, i.e. organically bound C-14. A comparison exercise was organised to identify suitable C-14 reference material. The aim was also to assess the statistical differences between techniques, levels of precision, sensitivity, practical advantages and disadvantages of each technique. C-14 is naturally occurring nuclide in environment, above ground nuclear test of 1955 and 1963 increased the amount of C-14 in biosphere and atmosphere. Four different materials were chosen for the comparison, three IAEA carbon series reference material and one mix of environmental biota sample. Homogeneity and possible degradation of the standards with time was checked analysing the C-13 content using a mass spectrometry technique. Four different methods used were direct counting, wet oxidation, combustion furnace and combustion coupled to conversion to benzene. Zeta and z-scores were used to identify discrepant laboratories. Combustion coupled to conversion to benzene is very specialised and gives a very low limit of detection compared with other techniques Combustion technique gives second best limit of detection and has the advantage of allowing simultaneous C-14 and H-3 measurements. All techniques successfully analysed the C-14 content. Liquid scintillation counter used significantly impacted on the limit of detection achieved. The biota sample has been demonstrated to be a viable QC material. A shortage of certified reference materials exists impacting on the ability to demonstrate that the techniques are fit for purpose. A participant commented that some of the backgrounds were higher than expected, and that possibly some background reducing options, such as low level count mode or high sensitivity count mode, may not have been used. A question regarding the standardisation of the vial was asked as C-14 sticks to glass vials. Jane answered that the laboratories were asked to do the measurements the same way they would do their routine analysis, so no parameters were fixed. 3. Invited talk: The TDCR method in LSC Philippe Cassette, CEA, France The TCDR technique is an absolute measurement method which doesn't require the use of a standard to calculate the activity. The model used to calculate the detection efficiency the Free parameter model - is based on the statistical description of the light emission. The mean number of photons emitted after absorption of an energy E by the liquid scintillator can be
3 calculated using the semi-empirical formula given by Birks, which is dependent on E, the electron stopping power and `Birks parameter' kb. The ionisation quench effect is stronger for radiation with high stopping power as recombination effect becomes more probable than radiative transfer. In a system with three photomultiplier tubes (PMTs), the free parameter model is calculated using the Triple to Double Coincidence Ratio (TDCR). In a real TDCR system, the three PMTs are all different, these differences have to be taken into account in the calculations. The kb parameter can be deduced by experimentation as for the correct kb value the calculated activity should always remain the same with decreasing efficiency. Monte Carlo simulation is the method used to calculate the uncertainty on the measurement. In the 300SL commercial TDCR counter from Hidex, the TDCR value is only used as a quench indicator. Measurements are being carried out at the moment to see if it could be used for the full TDCR method. The TDCR method is a mature LSC standardisation technique which is suitable for the standardisation of pure beta and some electron-capture emitters. For the moment, the use of the TDCR technique is limited to in-house build three PMT counters, but this could change if the commercially available TDCR counters are found to be suitable for the application of this technique. The next international LSC conference will be organised in Paris from the 6-10 September A participant asked if this method could be used for alpha emitters. Philippe commented that the TDCR method wasn't necessary for alpha emitters as the efficiency for an unquenched sample is almost one. Only a small correction of a few tenths of a percent is required because of the wall effect. A participant asked what size of PMT is used? In the traditional counter, two inch diameter Burle/RCA 8850 PMTs are used which are high gain with the first dynode of this PMT coated with gallium phosphide giving a very high emission of the first dynode and a good resolution on the single photon electron peak. 4. Invited talk: Necessary Quality Control for LSC Cocktail Production Ronald Edler, PerkinElmer The PerkinElmer Groningen site has three production groups: Microplates, luminescence reagents and LSC cocktails and supplies. The quality control of LSC cocktails starts with checking that the raw material is within acceptable parameters (colour, LSC background, luminescence, efficiency..). On some LSC cocktails, the short expiry date is quite often due to the fact that the luminescence reaches an unacceptable level after 2 to 3 years. K-40 contamination of nonylphenolethoxylates (NPE), an important component of LSC cocktails, is an issue at the moment as low-cost potassium catalysts are used in the production of NPEs. A reformulation of the LSC cocktails is necessary as a result of the environmental ban on the use of NPE which are endocrine disruptors in the aquatic environment. Alcohol ethoxylates are considered as an alternative to NPEs, but they are quite often prepared from bio-chemicals which contain a higher concentration of C-14. Changes in legislation and the world chemical market requires continuing support and efforts in LSC counting. As a result of the increasing K-40 and C-14 content in the raw materials, PerkinElmer has decided to create a new LS cockail for low level activity measurements, Ultima Gold u-llt which guarantees the presence of no K-40 and C-14 from bio-chemical sources in this cocktail.
4 A participant asked if it was possible to use the new Ultima Gold u-llt cocktail with an oxidiser. Ronald thinks it is possible in principle to use the new cocktail with oxidisers, but the use of a co-solvent to increase the uptake for C-14 is necessary. The reason for changing the composition of PicoFluor 15 with non-npe chemicals was questioned as it is already not a safe cocktail. It was replied that NPEs are banned from many countries, and as a consequence of the ban the production levels have dropped and some NPEs are produced only once a year. This creates some logistics issues as it required storing enough NPE from the production of specific cocktails for a whole year. A participant inquired as to why transition away from NPEs started with PicoFluor 15 rather than Ultima Gold. The reply given was that different types of NPEs are required for different cocktails and in the case of PicoFluor 15 the NPE wasn't produced anymore. It was enquired if there was a policy on cocktail names when the chemical composition of the cocktail changes. Usually there is an additional letter, e.g. UG LLT becomes UG u-llt but if the change is not significant then no modification is done to the name. A participant further suggested to change the name of the cocktail when the atomic composition changes as it can disturb measurements. However, unfortunately inter-batch changes are unavoidable due to the use of industrial products in the fabrication process which inevitably vary from one shipment to the next. The aim is to keep the efficiency, background and luminescence levels the same from one batch to the next rather than the atomic composition. 5. Invited talk: MCERTS for radioanalysis of waters Rob Allott, Environment Agency MCERTS is the Environment Agency's Monitoring CERTification Scheme. The aim is to promote the production of quality monitoring data. The sampling and chemical testing of the water, principally effluents, is a part of the MCERTS scope. This new scheme was necessary as the water quality industry is moving towards operator self-monitoring of their effluents. The various benefits of the MCERTS include the reliability of the data, gives clear guidance on what performance level is expected, indicates importance of quality in testing, provides confidence in the system of operator self-monitoring and limits the number of independent check monitoring. The MCERTS for radio-analysis of water will apply to contractors working for the Environment Agency and to Nuclear sites and their contractors. This MCERT is based on ISO 17025, but contains more details regarding the selection and validation of methods, performance targets (bias, precision), analytical quality control and reporting. The key issues are: Which radionuclides and groups to include in the standard, method validation, decision threshold and detection limits, performance criteria. Once the MCERT standard is finalised it will be available as an electronic copy on the Environment Agency website. The accreditation to ISO will be necessary in order to have accreditation to the MCERTS. A participant asked if it this new MCERT standard will be prescriptive similarly to the UKAS standard. Rob replied that only the method validation is prescriptive otherwise one could choose whatever method they wish. A participant enquired into the timescales. Originally the first draft was expected to be ready by end 2009, but it had been delayed.
5 6. talk: Approaches to measurement uncertainty evaluation Peter Harris Peter provided a list of useful documents for the evaluation of measurement uncertainty is given: GUM from BIPM, M3003 from UKAS, GPG 11 and SSfM documents from (see slides for more details). Uncertainty of measurement is a quantitative statement of one s knowledge of the quality of a measured value. Uncertainties can come from the item being measured, the measuring instrument and the environment. To evaluate the uncertainty on a measurement the knowledge of the input quantities and the model of the measurement are required. The model on the measurement can be a mathematical formula, a computer program or a numerical procedure that takes the values of the input quantities and delivers a value of the output quantity. Probability distributions are defined by their expectation and standard deviation. Examples of distribution include rectangular, Gaussian and the t-distribution. Type A uncertainty refers to uncertainty on the input quantity that is evaluated by statistical analysis of a series of values and type B refers to any other approach of evaluating the uncertainty. Uncertainty propagation can be done via analytical, conventional and numerical methods. Monte Carlo is a numerical method based on propagation of discrete representations of the distributions for the input quantities through the measurement model. In the Bayesian approach, there is a distinction between the measured quantities and the quantities about which prior information is available. This method usually requires the use of numerical methods. Are there any software programs to help implement the GUM? There are many software packages out there such as GUM workbench, BMTA spreadsheet for conventional uncertainty calculation, Add-ons to Excel for Monte Carlo calculations. 7. Any other business Glenis Tellett introduced the new Knowledge and Innovation centre which provides business accommodation for companies who rely on measurement for competitive advantage. Companies will get access to 's expertise, measurement and test facilities. Alan DuSautoy explained to the participants how the Radioactivity project formulation will be produced in future to improve impact and be more responsive to the UK needs by having a rolling programme formulation that will be carried out each year. Any feedback or new ideas on the programme formulation are welcome. has planned to organise a liquid scintillation training course and aims to hold it at the end of 2009 or early in Arzu Arinc Secretary, LSUF
The TDCR method in LSC. P. Cassette Laboratoire National Henri Becquerel CEA/LNE, France
The TDCR method in LSC P. Cassette Laboratoire National Henri Becquerel CEA/LNE, France LIQUID SCINTILLATION USERS FORUM 2009 Summary I. Some information on LSC II. LSC in metrology: the free parameter
More informationValidation of detection efficiency of alpha particles in commercial liquid scintillation counters
Validation of detection efficiency of alpha particles in commercial liquid scintillation counters Liquid Scintillation Users' Forum 14th October 2008 Andy Pearce Radionuclide and Neutron Metrology Group
More informationTDCR in a nutshell. P. Cassette, Laboratoire National Henri Becquerel, France
TDCR in a nutshell P. Cassette, Laboratoire National Henri Becquerel, France Summary LSC in radionuclide metrology, free parameter model The TDCR model Examples of relations between efficiency and TDCR
More informationHidex 300 SL Automatic TDCR Liquid Scintillation Counter
Hidex 300 SL Automatic TDCR Liquid Scintillation Counter The new dimension of versatility Our mission is to help your lab become more effective and make your work a pleasure by providing user-friendly
More information1220 QUANTULUS The Ultra Low Level Liquid Scintillation Spectrometer
1220 QUANTULUS The Ultra Low Level Liquid Scintillation Spectrometer PerkinElmer LAS (UK) Ltd, Chalfont Rd, Seer Green, Beaconsfield, Bucks HP9 2FX tel: 0800 896046 www.perkinelmer.com John Davies January
More information19:00 21:30 Registration and reception at Hotel ETAP Altinel. Welcome
AGENDA FOR 5th VERMI YOUNG RESEARCHERS WORKSHOP ON STANDARDISATION OF S in the frame of IPA Turkey* 1 6 November 2009, TAEK, Ankara, Turkey Sunday, 1 November 2009 19:00 21:30 Registration and reception
More informationFundamentals of Radionuclide Metrology
Fundamentals of Radionuclide Metrology Brian E. Zimmerman, PhD Physical Measurement Laboratory National Institute of Standards and Technology Gaithersburg, MD USA SIM Metrology Workshop Buenos Aires, Argentina
More informationAcoustics and Ionising Radiation Formulation and Strategy. 13 November 2008 Alan DuSautoy
Acoustics and Ionising Radiation Formulation and Strategy 13 November 2008 Alan DuSautoy Contents What is the future of Programme Formulation? What is Rolling Formulation? Programme Overview Roadmaps Future
More informationTime-Resolved Liquid Scintillation Counting
CHAPTER 44 Time-Resolved Liquid Scintillation Counting Norbert Roessler, Robert J. Valenta, and Stat van Cauter ABSTRACT A comparison is made between standard, two-tube coincidence liquid scintillation
More information1/5 Specifications Hidex 300 SLL automatic TDCR liquid scintillation counter
1/5 Specification Hidex 300 SL-SLL 100-240V AC and 24 V DC. Features Model (425-020) Super Low Level (SLL) Automatic TDCR Liquid Scintillation Counter with Guard Detector and Low Level PMTs General Description
More informationBasics of a/b Discrimination for Liquid Scintillation Counting
application Note Liquid Scintillation Basics of a/b Discrimination for Liquid Scintillation Counting Author Ronald Edler, PhD and Chuck Passo PerkinElmer, Inc. 940 Winter Street Waltham, MA USA Introduction
More informationHidex LSCs. LSC 2013 Barcelona
4.9.2013 1 4.9.2013 2 Hidex LSCs LSC 2013 Barcelona On Sunday March 24th Hidex is celebrating 20 years in the liquid scintillation counting community. Method to validate beef needed? Contents Hidex 300
More informationBUREAU INTERNATIONAL DES POIDS ET MESURES
1 BUREAU INTERNATIONAL DES POIDS ET MESURES International comparison of activity measurements of a solution of 3 H (January 2009) Participating laboratory: T ½ = (4 496.862 d; u = 9.131 d)* Ampoule number
More informationCalibration of Surface Contamination Monitors. UKAS Accredited Organisations Comparison. Final Report
Calibration of Surface Contamination Monitors UKAS Accredited Organisations 2007 Comparison Final Report 1 Normal expectations (ISO17025): Comparison participation at least annually Initial analysis of
More informationLS6500 Scintillation Counting Systems
LS6500 Scintillation Counting Systems Agenda LS6500 Concepts and Design H# Plus - The Basis of Advanced Technology Low Level Counting Alpha/Beta Discrimination LS6500 Concepts Give reproducible results
More informationTIME-RESOLVED LIQUID SCINTILLATION COUNTING
[RADOCARBON, VOL 32, No. 3, 1990, P 381-386] TME-RESOLVED LQUD SCNTLLATON COUNTNG MCHAEL KESSLER Packard nstrument Company, One State Street, Meriden, Connecticut 06450 ABSTRACT. Historically, scientists
More informationEnergy Calibration of Liquid Scintillation Counter to allow Semi-Qualitative Nuclide Identification in Water Samples
Energy Calibration of Liquid Scintillation Counter to allow Semi-Qualitative Nuclide Identification in Water Samples Mashaba, Machel. 1*, Kotze, Deon. 2, Tshivhase, Victor.M. 1, Faanhof, Arnaud. 1, 2 1
More informationConsistency and confidence: the role of the National Measurement System
Consistency and confidence: the role of the National Measurement System Dr. Martyn Sené Deputy Director, NPL 30 th June, 2011 Welcome to the National Physical Laboratory Consistent measurement - important
More informationTECHNETIUM-99 IN WATER
Analytical Procedure TECHNETIUM-99 IN WATER (TEVA DISC METHOD) 1. SCOPE 1.1. This procedure describes a method to separate and measure technetium-99 in water. 1.2. This method does not address all aspects
More informationWorking Copy. P /ru /,6. 8-tt tt. EffectiveDate: I ^, Operation and Calibration of the Tri-Carb Liquid Sclntillation Counter QAM-RI.106.
QAM-RI.106 Operation and Calibration of the Tri-Carb Liquid Sclntillation Counter Revision: 0 Laboratory Man ager/lqao/rso Date Date EffectiveDate: I ^, P /ru /,6 8-tt tt Renewal Date: lnitials: Texas
More informationLiquid Scintillation at NPL: Practical Applications
2003 LS User s Forum Liquid Scintillation at NPL: Practical Applications Andy Pearce CAIR, NPL 2 nd April 2003 Introduction Overview of applications Current projects Neutron-activated thioacetamide discs
More informationGross Alpha-Gross Beta Analysis in Water by Liquid Scintillation Counting (LSC)
Gross Alpha-Gross Beta Analysis in Water by Liquid Scintillation Counting (LSC) Bob Read, Ph.D. Director, Environmental Chemistry Laboratory Tennessee Department of Health Division of Laboratory Services
More informationTDCR and CIEMAT/NIST Liquid Scintillation Methods applied to the Radionuclide Metrology
TDCR and CIEMAT/NIST Liquid Scintillation Methods applied to the Radionuclide Metrology Paulo A. L. da Cruz 1, Carlos J. da Silva 1, Akira Iwahara 1, Jamir S. Loureiro 1, Antônio E. de Oliveira 1, Luiz
More informationThis document is a preview generated by EVS
INTERNATIONAL STANDARD ISO 13168 First edition 2015-07-01 Water quality Simultaneous determination of tritium and carbon 14 activities Test method using liquid scintillation counting Qualité de l eau Détermination
More informationAPPLICATION OF A FREE PARAMETER MODEL TO PLASTIC SCINTILLATION SAMPLES
APPLICATION OF A FR PARAMTR MODL TO PLASTIC SCINTILLATION SAMPLS Alex Tarancón Sanz 1, Hector Bagan 1, Karsten Kossert 2, Ole Nähle 2 1 Departmento de Química Analitica de la Universidad de Barcelona.
More informationDepth Distribution of H-3, C-14 and Co-60 in Decommissioning of the Biological Shielding Concrete of KRR-2
Depth Distribution of H-3, C-14 and Co-60 in Decommissioning of the Biological Shielding Concrete of KRR-2 S.B. Hong, H.R. Kim, K.H. Chung, K.H. Chung, J.H. Park Korea Atomic Energy Research Institute
More informationInternational Atomic Energy Agency. Department of Nuclear Sciences and Applications. IAEA Environment Laboratories
International Atomic Energy Agency Department of Nuclear Sciences and Applications IAEA Environment Laboratories Vienna International Centre, P.O. Box 100, 1400 Vienna, Austria REFERENCE SHEET CERTIFIED
More informationTECHNETIUM-99 IN SOIL
Analytical Procedure TECHNETIUM-99 IN SOIL 1. SCOPE 1.1. This procedure describes a method to separate and measure technetium-99 in soil. 1.2. This method does not address all aspects of safety, quality
More informationNEW DEVELOPMENTS OF AUTORADIOGRAPHY TECHNIQUE TO IMPROVE ALPHA AND BETA MEASUREMENTS FOR DECOMMISSIONING FACILITIES
NEW DEVELOPMENTS OF AUTORADIOGRAPHY TECHNIQUE TO IMPROVE ALPHA AND BETA MEASUREMENTS FOR DECOMMISSIONING FACILITIES PASCAL FICHET, C MOUGEL, Y DESNOYERS, P SARDINI, H WORD DEN SERVICE D ETUDES ANALYTIQUES
More informationWork Programme. of the International Bureau of Weights and Measures. for the four years Comité international des poids et mesures
Work Programme of the International Bureau of Weights and Measures for the four years 2016-2019 Comité international des poids et mesures Section II : BIPM Work Programme for 2016-2019 35 Priority activities
More informationČerenkov counting and liquid scintillation counting of 36 Cl
Čerenkov counting and liquid scintillation counting of 36 Cl Karsten Kossert, Ole Nähle Physikalisch-Technische Bundesanstalt (PTB), Braunschweig, Germany and Agustín Grau Carles Instituto de Física Fundamental
More informationLIQUID SCINTILLATION COUNTERS. {Beta Counters}
LIQUID SCINTILLATION COUNTERS {Beta Counters} We offer a range of Alpha, Beta & Gama counters, from Hidex Oy, Finland to meet your specific Liquid Scintillation counting requirements. Triathler LSC Sense
More informationInternational Atomic Energy Agency. Department of Nuclear Sciences and Applications. IAEA Environment Laboratories
International Atomic Energy Agency Department of Nuclear Sciences and Applications IAEA Environment Laboratories Vienna International Centre, P.O. Box 100, 1400 Vienna, Austria REFERENCE SHEET CERTIFIED
More informationJ. Keightley 1 P. Cassette 2, L. Johansson 1
RADIONUCLIDE METROLOGY FOR NEW GENERATION NUCLEAR POWER PLANTS J. Keightley 1 P. Cassette 2, L. Johansson 1 1 NPL, Teddington, United Kingdom 2 CEA-LNHB, Saclay, France E-mail (corresponding author): john.keightley@npl.co.uk
More informationHidex Liquid Scintillation Counters
Hidex Liquid Scintillation Counters The new dimension of versatility Our mission is to help your lab become more effective and make your work a pleasure by providing user-friendly instruments with ultimate
More informationThe measurement of tritium in environmental water samples
The measurement of tritium in environmental water samples 1. Introduction The Hidex 300 SL automatic TDCR liquid scintillation provides an excellent solution for the determination of H-3 in drinking water.
More informationStandardization of Tritium by CIEMAT/NIST Method and TDCR Method
Standardization of Tritium by CIEMAT/NIST Method and TDCR Method Wu Yongle 1,3 ; Liu Haoran 1,2 ; Liang Juncheng 2 ; Liu Jiacheng 2 ; Yue Huiguo 3 ; Liu Senlin 1 ; Yang Yuandi 2 ; Yuan Daqing 1 ; 1.China
More informationTECHNETIUM-99 IN WATER
Analytical Procedure TECHNETIUM-99 IN WATER (WITH VACUUM BOX SYSTEM) 1. SCOPE 1.1. This procedure describes a method to separate and measure technetium-99 in water. 1.2. This method does not address all
More informationActivity measurement of 55 Fe within the scope of the BIPM comparison 2006
Activity measurement of 55 Fe within the scope of the BIPM comparison 2006 Karsten Kossert PTB, Department 6.1 Radioactivity BIPM Workshop 2 on CCRI(II) Activity Uncertainties and Comparisons, 17-18 September
More informationLiquid Scintillation Counter
Instrumentation & Methods: s & Tritium Richard Sheibley Pennsylvania Dept of Env Protection Principle Beta particle emission Energy transferred to Solute Energy released as UV Pulse Intensity proportional
More informationRESULTS YOU CAN COUNT ON
RESULTS YOU CAN COUNT ON Liquid Scintillation Counting Principle and the Application in Biological Research 1 2009 PerkinElmer liquid scintillation counting β particle Tri-Carb family of liquid scintillation
More informationActivity determination of 88 Y by means of 4πβ(LS)-γ coincidence counting
Activity determination of 88 Y by means of 4πβ(LS)-γ coincidence counting Justyna Marganiec-Galazka Ole J. Nähle Karsten Kossert Division 6 Ionizing Radiation Department 6.1 Radioactivity Working Group
More informationISO INTERNATIONAL STANDARD
INTERNATIONAL STANDARD ISO 11704 First edition 2010-07-01 Water quality Measurement of gross alpha and beta activity concentration in non-saline water Liquid scintillation counting method Qualité de l'eau
More informationDETERMINING FUNDAMENTAL PARAMETERS OF A SINGLE-PHOTOTUBE LIQUID SCINTILLATION COUNTER
DETERMINING FUNDAMENTAL PARAMETERS OF A SINGLE-PHOTOTUBE LIQUID SCINTILLATION COUNTER Pall Theodórsson Science Institute, University of Iceland, Dunhaga 3, IS-107 Reykjavík, Iceland. Email: pth@raunvis.hi.is.
More informationISO Water quality Determination of carbon 14 activity Liquid scintillation counting method
INTERNATIONAL STANDARD ISO 13162 First edition 2011-11-01 Water quality Determination of carbon 14 activity Liquid scintillation counting method Qualité de l eau Détermination de l activité volumique du
More informationA Comparison of True Alpha Activities in Air Filter Samples with Values Obtained from Radioactivity-in-air Monitors
A Comparison of True Alpha Activities in Air Filter Samples with Values Obtained from Radioactivity-in-air Monitors Julian Dean 1, Maria Garcia Miranda 1 and Pete Burgess 2 1 National Physical Laboratory,
More informationClearance Monitoring. Chris Goddard.
Clearance Monitoring Chris Goddard Outline What is Clearance? Clearance Limits around Europe Measurement techniques Plastic scintillators Long Range Alpha Detection Example systems Thermo SAM12 VF FRM-2
More informationInvestigation of Uncertainty Sources in the Determination of Gamma Emitting Radionuclides in the WBC
Investigation of Uncertainty Sources in the Determination of Gamma Emitting Radionuclides in the WBC A. Specification Whole body counting method is used to detect the gamma rays emitted by radio nuclides,
More informationCOMPARATIVE STUDY BETWEEN EXTERNAL STANDARD METHOD AND INTERNAL STANDARD METHOD FOR LOW-LEVEL TRITIUM MEASUREMENTS
International Conference Nuclear Energy in Central Europe 2001 Hoteli Bernardin, Portorož, Slovenia, September 10-13, 2001 www: http://www.drustvo-js.si/port2001/ e-mail: PORT2001@ijs.si tel.:+ 386 1 588
More informationAcronyms, Abbreviations, and Symbols Foreword to the First Edition Foreword to the Second Edition Preface to the First Edition Preface to the Second
Contributors p. xxix Acronyms, Abbreviations, and Symbols p. xxxi Foreword to the First Edition p. xliii Foreword to the Second Edition p. xlv Preface to the First Edition p. xlvii Preface to the Second
More informationC2-05: Creation of national standards for some emerging pharmaceutical radionuclides to ensure the radioprotection of patients and medical staffs
C2-05: Creation of national standards for some emerging pharmaceutical radionuclides to ensure the radioprotection of patients and medical staffs Project Leaders: IFIN-HH/DRMR/LMR: Dr. Aurelian Luca ;
More informationStrontium 90 in Urine: Evaluation of Radiotoxicological Techniques by International PROCORAD Comparison Exercises
Strontium 90 in Urine: Evaluation of Radiotoxicological Techniques by International PROCORAD Comparison Exercises R Fottorino 1, M. Ruffin 1 and Ph. Bérard 2 1 CEA Cadarache, LABM, Bâtiment 102, 13108
More information12/18/2016. Radioanalysis Laboratory Capabilities and Issues. Eleventh Annual Radiation Measurement Cross Calibration Workshop RMCC XI
1. Introduction 2. Laboratories Gamma spectrometry laboratory. Radiochemistry laboratory. Analytical chemistry laboratory. Alpha spectrometry laboratory Gross alpha / beta laboratory. Neutron activation
More informationAutomation of the radiochemical procedures for the sequential separation of radionuclides
LSC2017 - An International Conference on Advances in Liquid Scintillation Spectrometry, Copenhagen Denmark, 1 5 May 2017 Automation of the radiochemical procedures for the sequential separation of radionuclides
More informationRadio-HPLC Flow Cells
Radio-HPLC Flow Cells Innovative Technology Expands the Limits of Radio-HPLC Detection The Most Versatile Flow Cells Packard s Radiomatic TM flow scintillation analyzers and flow cells have been integrated
More informationSLAC Radioanalysis Laboratory
SLAC Radioanalysis Laboratory Henry Brogonia Dosimetry and Radiological Protection Group (DREP) DOE Environmental Radiation Protection Program Review (July 23-24, 2007) Radioanalysis Laboratory Mission
More informationMeasurement of Tritium in Helium
detect and identify Measurement of Tritium in Helium Dr. Alfred Klett Berthold Technologies, Bad Wildbad, Germany 22 nd Annual Air Monitoring Users Group (AMUG) Meeting Palace Station Hotel, Las Vegas,
More informationISO Water quality Strontium 90 and strontium 89 Test methods using liquid scintillation counting or proportional counting
INTERNATIONAL STANDARD ISO 13160 First edition 2012-07-15 Water quality Strontium 90 and strontium 89 Test methods using liquid scintillation counting or proportional counting Qualité de l eau Strontium
More informationLSC for Quality Control of 99m TC Eluate from 99 Mo- 99m Tc Generator
LSC2017 Conference 1-5th May, 2017, Copenhagen LSC for Quality Control of 99m TC Eluate from 99 Mo- 99m Tc Generator Xiaolin Hou Technical University of Denmark, Center for Nuclear Technologies Roskilde,
More informationISO INTERNATIONAL STANDARD. Measurement of radioactivity in the environment Soil Part 5: Measurement of strontium 90
INTERNATIONAL STANDARD ISO 18589-5 First edition 2009-03-01 Measurement of radioactivity in the environment Soil Part 5: Measurement of strontium 90 Mesurage de la radioactivité dans l'environnement Sol
More informationNorthwest University, Mafikeng, Centre of Applied Radiation Science & Technology (CARST), South Africa.
Gross Alpha and Beta Measurements of Water Samples from the Wonderfonteinspruit Catchment Area in the Gauteng Province (South Africa), using Liquid Scintillation Counting Machel Mashaba 1, Deon Kotze 2,
More information2100TR Liquid Scintillation Counter
2100TR Liquid Scintillation Counter Description The Tri-Carb 2100TR liquid scintillation counter is computer-controlled, bench top liquid scintillation analyzer for detecting small amounts of alpha, beta
More informationRadiation Safety Protection for Callahan Eye Hospital (OHS_RS502)
Introduction Welcome to the Radiation Safety Protection for Callahan Eye Hospital Training Course (OHS_RS502). This training is designed and required for anyone working with or around Radioactive Materials
More informationManufacturing and Measurement of photon-emitting Wide Area Reference Sources according to ISO 8769
Manufacturing and Measurement of photon-emitting Wide Area Reference Sources according to ISO 8769 Klaus Thieme Eckert & Ziegler Nuclitec GmbH Braunschweig 1 Content Introduction Measurements and Results
More informationDETERMINATION OF CORRECTION FACTORS RELATED TO THE MANGANESE SULPHATE BATH TECHNIQUE
DETERMINATION OF CORRECTION FACTORS RELATED TO THE MANGANESE SULPHATE BATH TECHNIQUE Ján Haščík, Branislav Vrban, Jakub Lüley, Štefan Čerba, Filip Osuský, Vladimír Nečas Slovak University of Technology
More informationUnit 2. Instrumentation. Experts Teaching from Practical Experience
Unit 2 Instrumentation Experts Teaching from Practical Experience Gas-Filled Detectors Gas-filled detectors measure the charge released when radiation interacts with the gas Three types: Ion Chambers,
More informationPete Burgess, Nuvia Limited. Clearance and exemption
Pete Burgess, Nuvia Limited Clearance and exemption The clearance, exclusion and exemption process Most of the UK nuclear industry (and many other organisations) refer to the Clearance and Exemption Working
More informationPractical Approaches using TDCR Measurements and Alpha/Beta Separation
Practical Approaches using TDCR Measurements and Alpha/Beta Separation Jost Eikenberg, Maya Jäggi, Andreas Brand Division for Radiation Protection and Safety Paul Scherrer Institute, CH-5232 Villigen Overview
More informationCIEMAT/NIST STANDARDIZATION METHOD EXTENDED TO ANODE OUTPUTS FOR BETA AND ELECTRON-CAPTURE NUCLIDES J. F. ORTIZ
CIEMAT/NIST STANDARDIZATION METHOD EXTENDED TO ANODE OUTPUTS FOR BETA AND ELECTRON-CAPTURE NUCLIDES J. F. ORTIZ Departmento de Mecanica, Escuela Tecnica Superior de Ingenieros Industriales (ETSII), Universidad
More informationEN-TRAP. The European Network of Testing Facilities for the Quality Checking of Radioactive Waste Packages
EN-TRAP The European Network of Testing Facilities for the Quality Checking of Radioactive Waste Packages Pierre Van Iseghem SCK CEN 1 March 2008 Why / when characterize RW? Independent Verification Regulator
More informationDisproof of solar influence on the decay rates of 90 Sr/ 90 Y
Disproof of solar influence on the decay rates of 90 Sr/ 90 Y Karsten Kossert * and Ole J. Nähle Physikalisch-Technische Bundesanstalt (PTB), Bundesallee 100, 38116 Braunschweig, Germany Abstract A custom-built
More informationPrimary Standardization of 152 Eu by 4πβ(LS) γ (NaI) coincidence counting and CIEMAT-NIST method
Primary Standardization of 152 Eu by 4πβ(LS) γ (NaI) coincidence counting and CIEMAT-NIST method A Ruzzarin 1, P A L da Cruz 2, A L Ferreira Filho 2, A Iwahara 2 1 Laboratório de Instrumentação Nuclear/Programa
More informationISO Water quality Measurement of polonium 210 activity concentration in water by alpha spectrometry
INTERNATIONAL STANDARD ISO 13161 First edition 2011-10-01 Water quality Measurement of polonium 210 activity concentration in water by alpha spectrometry Qualité de l eau Mesurage de l activité du polonium
More informationA New Measurement Method For Tritium Rapid Detection by LSC. IM.CAEP Xu Qinghua
A New Measurement Method For Tritium Rapid Detection by LSC IM.CAEP Xu Qinghua xuqinghuawork@sina.com Outline Background Results and discussion Materials and methods Background FOM value for channel Sample
More informationStandardization of 18 F by Digital β(ls)-γ Coincidence Counting
Standardization of 18 F by Digital β(ls)-γ Coincidence Counting Rodrigues D. 1, Balpardo C. 1, Cassette P. 2, Arenillas P. 1, Capoulat M. E. 1, Ceruti G. 1, García-Toraño E. 3 1 Laboratorio de Metrología
More informationTechnical note on using JEFF-3.1 and JEFF data to calculate neutron emission from spontaneous fission and (α,n) reactions with FISPIN.
Page 1 of 11 Technical note on using JEFF-3.1 and JEFF-3.1.1 data to calculate neutron emission from spontaneous fission and (α,n) reactions with FISPIN. Nexia Solutions Ltd Dr. Robert W. Mills and Dr.
More informationLiquid Scintillation Counting Performance Using Glass Vials in the Wallac 1220 QuantulusTM
CHAPTER 43 Liquid Scintillation Counting Performance Using Glass Vials in the Wallac 1220 QuantulusTM Lauri Kaihola ABSTRACT Low-potassium glass vials can obtain reduced background count rates for beta
More informationBasic hands-on gamma calibration for low activity environmental levels
Basic hands-on gamma calibration for low activity environmental levels Iolanda Osvath Presented by Mats Eriksson Environment Laboratories Marine Environment Laboratories, Monaco Radiometrics Laboratory
More informationHalf-life data a critical review of TECDOC-619 update
ARTICLE IN PRESS Applied Radiation and Isotopes 60 (2004) 257 262 Half-life data a critical review of TECDOC-619 update M.J. Woods a, *, S.M. Collins b a Ionising Radiation Metrology Consultants Ltd, 152
More informationInvestigation of Uncertainty Sources in the Determination of Gamma Emitting Radionuclides in the UAL
Investigation of Uncertainty Sources in the Determination of Gamma Emitting Radionuclides in the UAL A. Specification Gamma-spectrometry method is used to identify and determine the activity concentration
More informationInternational Conference on the Safety of Radioactive Waste Management
IAEA - CN - 78 / 43 International Conference on the Safety of Radioactive Waste Management 13-17 March 2000, Cordoba, Spain Subject : Assessment of the safety of radioactive waste management Considerations
More informationRadiation Detection. 15 th Annual OSC Readiness Training Program.
Radiation Detection 15 th Annual OSC Readiness Training Program www.oscreadiness.org GM Detectors 15 th Annual OSC Readiness Training Program www.oscreadiness.org 1 A closer look 15 th Annual OSC Readiness
More informationUnit 4. Statistics, Detection Limits and Uncertainty. Experts Teaching from Practical Experience
Unit 4 Statistics, Detection Limits and Uncertainty Experts Teaching from Practical Experience Unit 4 Topics Statistical Analysis Detection Limits Decision thresholds & detection levels Instrument Detection
More informationUncertainty of Radionuclide Calibrator Measurements of 99
Uncertainty of Radionuclide Calibrator Measurements of 99Tc m 99 Tc I.S.Hufton, Department of Medical Physics, Royal Liverpool University Hospital NPL Intercomparison Exercise Traceability exercise Multi-centre
More informationMonitoring of Ionizing Radiations facilities Experience and challenge
Monitoring of Ionizing Radiations facilities Experience and challenge Lebanese Atomic Energy Commission The LAEC was established in 1996 with the full support and assistance of the IAEA, having the mandate
More informationarxiv: v1 [physics.ins-det] 3 Feb 2011
Nuclear Instruments and Methods in Physics Research A 00 (2018) 1 5 Alogo.pdf Nuclear Instruments and Methods in Physics Research A Scintillation decay time and pulse shape discrimination in oxygenated
More informationGuidance for best practice from MetroMRT. Lena Johansson, NPL
Guidance for best practice from MetroMRT Lena Johansson, NPL National Measurement Institutes MetroMRT is an NMI led European project looking to improve and provide tools for clinical implementation of
More informationPrimary Standardisation of I-125
Primary Standardisation of S. Pommé, T. Altzitzoglou, R. Van Ammel, G. Sibbens Institute for Reference Materials and Measurements (IRMM) Geel, Belgium http://www.irmm.jrc.be http://www.jrc.cec.eu.int Stefaan
More informationInternational Atomic Energy Agency. Department of Nuclear Sciences and Applications. IAEA Environment Laboratories
International Atomic Energy Agency Department of Nuclear Sciences and Applications IAEA Environment Laboratories Vienna International Centre, P.O. Box 100, 1400 Vienna, Austria REFERENCE SHEET CERTIFIED
More informationAnnex to the Accreditation Certificate D-PL according to DIN EN ISO/IEC 17025:2005
Deutsche Akkreditierungsstelle GmbH Annex to the Accreditation Certificate D-PL-14595-01-00 according to DIN EN ISO/IEC 17025:2005 Period of validity: 22.12.2017 to 13.10.2021 Holder of certificate: Forschungszentrum
More information*Corresponding author,
Horia Hulubei National Institute of Physics and Nuclear Engineering g - IFIN HH Horia Hulubei National Institute of Physics and Nuclear Engineering, IFIN-HH ROMANIA *Corresponding author, e-mail bercea@nipne.ro
More informationPhys 243 Lab 7: Radioactive Half-life
Phys 243 Lab 7: Radioactive Half-life Dr. Robert MacDonald The King s University College Winter 2013 Abstract In today s lab you ll be measuring the half-life of barium-137, a radioactive isotope of barium.
More informationMetrology for Molecular Radiotherapy MetroMRT V. Smyth (coordinator) Lena Johansson, Andrew Fenwick, Kelley Ferreira et al.
EMRP Health(II) Metrology for Molecular Radiotherapy MetroMRT V. Smyth (coordinator) Lena Johansson, Andrew Fenwick, Kelley Ferreira et al. Metro-MRT Metrology for Molecular Radiotherapy, 2012-2015 WP
More informationAccreditation of radiochemical analyses, from NAMAS to ISO 17025:2005 and beyond
Accreditation of radiochemical analyses, from NAMAS to ISO 17025:2005 and beyond George Ham Centre for Radiation, Chemicals and Environmental Hazards Health Protection Agency The Analysts Dilemma: Maintaining
More informationPARTICLE MEASUREMENT IN CLEAN ROOM TECHNOLOGY
WHITEPAPER ENGLISH PARTICLE MEASUREMENT IN CLEAN ROOM TECHNOLOGY PARTICLE MEASUREMENT Particle measurement in cleanrooms. WP1508006-0100-EN, V1R0, 2015-08 PARTICLE MEASUREMENT IN CLEAN ROOM TECHNOLOGY
More informationCalibration & Use of a Capintec CAPTUS 3000 Portable Thyroid Uptake System for Iodine-125 Bioassay Measurements Todd W.
Calibration & Use of a Capintec CAPTUS 3000 Portable Thyroid Uptake System for Iodine-125 Bioassay Measurements Todd W. Baker, MSPH, CHP Photo image area measures 2 H x 6.93 W and can be masked by a collage
More informationDetectors for the measurement of ionizing radiation
For the measurement of radiation, the following reactions during the irradiation of matter are predominantly utilized: Ionization in gases (Ionization chamber, proportional flow counter, release counter)
More informationAntoni Borràs, Oscar Pozo, Victor Cerda, Laura Ferrer Universitat de les Illes Balears (Spain). Melissa Rodas Sciware Systems, Mallorca (Spain).
Design and development of a miniaturized detector for radionuclide determination in automated flow system Antoni Borràs, Oscar Pozo, Victor Cerda, Laura Ferrer Universitat de les Illes Balears (Spain).
More informationIonizing Radiation Metrology at NMIJ/AIST and ENEA-INMRI
Ionizing Radiation Metrology at NMIJ/AIST and ENEA-INMRI Pierino De Felice and Marco Capogni - ENEA C.R. Casaccia Istituto Nazionale di Metrologia delle Radiazioni Ionizzanti (ENEA-INMRI) Roma ITALY Yoshio
More informationAnalytical Technologies in Biotechnology Prof. Dr. Ashwani K. Sharma Department of Biotechnology Indian Institute of Technology, Roorkee
Analytical Technologies in Biotechnology Prof. Dr. Ashwani K. Sharma Department of Biotechnology Indian Institute of Technology, Roorkee Module - 2 Radioisotopes Techniques Lecture - 3 GM Counting and
More information